MPP PI Pinpointer probes - one of them has peculiar behavior - hyper sensitive to EF - ideas?

Collapse
X
 
  • Filter
  • Time
  • Show
Clear All
new posts

  • #16
    Have you tried increasing main sample delay suggested reply #8? The decay signal looks overdamped. Maybe sampling some X signal from the ferrite since it detects swinging ever so slowly.

    Comment


    • #17
      Originally posted by green View Post
      Have you tried increasing main sample delay suggested reply #8? The decay signal looks overdamped. Maybe sampling some X signal from the ferrite since it detects swinging ever so slowly.

      Not really. I can only adjust from 9 to 16us at the moment. What ballpark figure would be interesting to test for (20,40,80,160us) ?

      Comment


      • #18
        I am having intuition that the ferrite coil is saturating (or near to saturation), causing a decrease in inductance and invoking a transient overdamped condition.
        Remove some turns if you can, and see what happens.

        Comment


        • #19
          I am thinking that the ferrite you used is from some VHF type apllication and isn't suitable for lower freq. The pulse width and current is high enough to invoke some strange reaction of this ferrite.
          Increasing the main sample delay will defeat the original purpose of your pinpointer probe.

          Comment


          • #20
            Originally posted by dbanner View Post
            I am having intuition that the ferrite coil is saturating (or near to saturation), causing a decrease in inductance and invoking a transient overdamped condition.
            Remove some turns if you can, and see what happens.
            I agree, sounds like core saturation. You could also reduce the TX pulse width or add some series R to the coil.

            When I was at White's, the first pinpointer I built used a wrong ferrite and it turned out to be a very effective compass.

            Comment


            • #21
              Originally posted by Carl-NC View Post
              I agree, sounds like core saturation. You could also reduce the TX pulse width or add some series R to the coil.

              When I was at White's, the first pinpointer I built used a wrong ferrite and it turned out to be a very effective compass.
              Is there a way to measure core saturation? Maybe connect a 1 ohm resistor from +battery to coil and scope Tx on current(voltage across coil).

              Comment


              • #22
                Originally posted by green View Post
                Is there a way to measure core saturation?
                It's tricky, as you would really need to plot the B-H curve.
                Here's some food for thought:
                https://www.nutsvolts.com/magazine/a.../probing_cores

                Comment


                • #23
                  The inductance(ie number of turns) is too high for the Al value of this ferrite core PP2, causing it to saturate.
                  Reducing the number of turns would likely cure this. Or limiting the level of the pulse.

                  Comment


                  • #24
                    On second thought, perhaps am radio ferrite rod is best.

                    Comment


                    • #25
                      Originally posted by Qiaozhi View Post
                      It's tricky, as you would really need to plot the B-H curve.
                      Here's some food for thought:
                      https://www.nutsvolts.com/magazine/a.../probing_cores

                      Taken from the above article:
                      "typical core materials show a straight line which curves towards the horizontal at both ends (see Figure 1). The curved tips show the core's permeability falling at high currents." End quote

                      Green, you are welcome to build the B-H test circuit, as I know you are a TESTING freak-a-zoid.
                      Attached Files

                      Comment


                      • #26
                        Inductance: 170uH
                        Rprobe: 0.8 Ohms
                        Rseries to probe 56 Ohms
                        Windings: 64
                        Shielded: yes
                        probe specs from reply #1

                        charge TC=3us, peak current=.2A, coil current should be about .125A@3us, .173A@6us, .19A@9us. Was wondering if Polymer compared the current(voltage across a 1R resistor in series with the coil)with predicted if he couldn't tell if coil was saturating. If saturating, should get to .19A in less than 9us? Maybe charge wouldn't look exponential. If could paste recording in Excel it could be charted linear log to see if it charts a straight line.

                        Comment


                        • #27
                          Gosh, so many replies - Thank You!

                          One correction: Rseries to PP-Probe is 68 Ohms (not 56 Ohms) - sorry

                          Comment


                          • #28
                            Originally posted by green View Post
                            Inductance: 170uH
                            Rprobe: 0.8 Ohms
                            Rseries to probe 56 Ohms
                            Windings: 64
                            Shielded: yes
                            probe specs from reply #1

                            charge TC=3us, peak current=.2A, coil current should be about .125A@3us, .173A@6us, .19A@9us. Was wondering if Polymer compared the current(voltage across a 1R resistor in series with the coil)with predicted if he couldn't tell if coil was saturating. If saturating, should get to .19A in less than 9us? Maybe charge wouldn't look exponential. If could paste recording in Excel it could be charted linear log to see if it charts a straight line.


                            Heres a scope shot of the shunt voltage measured across the 68 Ohm series resistor of PP2: Click image for larger version

Name:	voltasgeviashunt_68ohms_PP2.jpg
Views:	1
Size:	75.2 KB
ID:	355507

                            Comment


                            • #29
                              Originally posted by dbanner View Post
                              The inductance(ie number of turns) is too high for the Al value of this ferrite core PP2, causing it to saturate.
                              Reducing the number of turns would likely cure this. Or limiting the level of the pulse.

                              It looks like you are right. I reduced the windings in steps of 5 and testing this "compass" (Thanks Carl) until I got to 35 windings less.
                              The problem is practically not noticable anymore. I also played with the sample delay - it did not make much difference.

                              The pre-amp output looks much different now, yet it still detects a nickel at around 3 inches.

                              Here's a scope shot: Click image for larger version

Name:	pp2minus35Windings_6us.png
Views:	1
Size:	27.1 KB
ID:	355508

                              Comment


                              • #30
                                Originally posted by Polymer View Post
                                It looks like you are right. I reduced the windings in steps of 5 and testing this "compass" (Thanks Carl) until I got to 35 windings less.
                                The problem is practically not noticable anymore. I also played with the sample delay - it did not make much difference.

                                The pre-amp output looks much different now, yet it still detects a nickel at around 3 inches.

                                Here's a scope shot: [ATTACH]48215[/ATTACH]

                                Comment