Increased Tx time to 16msec flat. Charted US quarter and 1 troy ounce 99.9% pure copper coin. Straight line lin log after 100usec. Why doesn't it continue straight below 100usec with a 16msec Tx? To find the TC the decay needs to be charted lin log. Maybe a scope linear trace would be just as good or better for looking at signal loss during target sample time. What would be a good time full scale? I'm thinking 50usec would cover most target sample times. Would 20usec or 100usec be better?Attached Files
Comment
-
Originally posted by green View PostIncreased Tx time to 16msec flat. Charted US quarter and 1 troy ounce 99.9% pure copper coin. Straight line lin log after 100usec. Why doesn't it continue straight below 100usec with a 16msec Tx? To find the TC the decay needs to be charted lin log. Maybe a scope linear trace would be just as good or better for looking at signal loss during target sample time. What would be a good time full scale? I'm thinking 50usec would cover most target sample times. Would 20usec or 100usec be better?
For a good explanation of this process, download the paper 'Applications of Transient Electromagnetic Techniques' Technical note TN-7 from the Geonics Ltd website. Page 6 for 'Target Responses'. Although this is for geophysical prospecting, the same scaled down principles apply for PI metal detectors.
Eric.
Comment
-
Originally posted by Ferric Toes View PostBy 'below' I assume you mean for times shorter than 100uS. This is because you are in the 'skin effect', 'early time', part of the decay curve. the eddy currents have not yet had time to diffuse to the core of the coin, due to it's high conductivity. At 100uS the diffusion ceases and then we have a single exponential; hence the straight line from that point on.
For a good explanation of this process, download the paper 'Applications of Transient Electromagnetic Techniques' Technical note TN-7 from the Geonics Ltd website. Page 6 for 'Target Responses'. Although this is for geophysical prospecting, the same scaled down principles apply for PI metal detectors.
Eric.
Comment
-
Here is the linear response to a silver ring 21mm O.D., 1mm thick and 4mm band. It showns how the amplitude peaks at 20uS delay and a TX width of 80uS. The measurements this instrument takes are that the TX pulse width increases by 4 x each delay figure. Hence at 10us delay the TX is 40uS. At 20uS delay the TX is 80uS and so on. All flat topping. The sample pulse width is equal to the delay for all measurements. This demonstrates that TX widths less than 80uS are below optimum for this ring.
Eric.
Comment
-
Charted ground decay with flat Tx, 160usec and 16000usec on time. Finally got a decay slope of -1. Looks like if Tx time is flat and long enough the slope is -1.Attached Files
Comment
-
Charted a silver dollar(1oz. fine silver) borrowed from a friend. My silver dollars were stolen in a break in awhile back. Close to the 1oz. 99.9% pure copper coin. Tx 16000usec flat.Attached Files
Comment
-
Originally posted by green View PostI was thinking after reading this thread, after 5 target TC's the eddy currents should have decayed with a flat Tx. 16000usec flat TX/5=3200usec TC. The copper coin with a TC of 500usec would have 32 TC's to decay during the flat TX. Maybe I'm missing something or doing something wrong. Does your Hocking have a different reading with stacked coins?
Eric.
Comment
-
Originally posted by green View PostCharted ground decay with flat Tx, 160usec and 16000usec on time. Finally got a decay slope of -1. Looks like if Tx time is flat and long enough the slope is -1.
Eric.
Comment
-
Originally posted by Ferric Toes View PostYour results are coming out good. Slope of t^-1 is what it should be. You will find very slight variations in ground from different places though. What is your method of plotting the results?
Eric.
Comment
-
Originally posted by Ferric Toes View PostI wonder if the simulated results are a bit different to what happens in practice. From what I have both read and observed in tests with a logarithmic front end, the response of a solid target such as a sphere or cube only settles to a single exponential
after one Tau. Prior to that, it is a sum of exponentials (the skins of an onion effect). This I believe can be simulated by having additional inductors and resistors mutually coupled to the single one.
This effect reduces for thin flat objects, such as coins or rings. A medium to thin gold ring will display a single exponential, while a coin such as a silver dollar will display the sum of exponentials effect.
Invariably PI detectors sample well with the Tau of many objects, particularly higher conductivity ones where the TX pulse is too short to excite the primary Tau. What we then we see is only the skin effect response and we wonder why a big silver dollar has much less range than a nickel.
Fundamental theory states that that for 99% excitation the TX pulse should be 5 x the target Tau although 3 x is still 95%. Likewise for maximum current and a almost flat termination of the field before switch off the TX pulse should be 5, or at least 3, time the coil TC. The easy way of improving the coil TC is to add some series resistance.
Eric.
Comment
-
Some scope pictures. Tx(160usec ramp, 1000pps) Target distance was adjusted for amplifier out=1.5 volts at 6usec delay. A shorter sample window be better for the 10x10mm can side. The copper coin has a faster decay than the US quarter in the target window with the 160usec Tx. Scope external trigger(fet gate command off).Attached Files
Comment
-
Originally posted by baum7154 View PostHello Green,
It would be interesting to see the 6mm x 6mm Al can side target in this test as well.
Regards,
Dan
Lost 6x6mm can side I tested before. Tested one cut this morning as close to 6x6mm as I could get.Attached Files